In the two preceding pieces we wrote that “you can just check the approval number”. This piece tells you the limits of that habit. Products intended solely for export also receive an MFDS marketing approval number. A number in exactly the same format as a domestic one, like “Je-heo 26-657”. And yet that product cannot be used on a patient in Korea. What separates the two is a single field on the lookup screen — “Export only” — and anyone can see it.

Clinical column About a 13-minute read September 2026 Miso Clinic, Daegu · Dr. Lee Chi-Hak

The conclusion, first

Export-only medical devices also receive a genuine marketing approval number. When we actually looked them up, a domestic product and an export-only product carried numbers in exactly the same format — “Je-heo 26-667” and “Je-heo 26-657”. What separates them is a single field inside the MFDS lookup screen: “Export only” (the field labelled 수출용에한함). And the scale of this is not small — of the 1,010 items returned by a search for “tissue-filling biomaterial”, 745 (73.8%) were export-only (looked up on 4 September 2026). Article 26(1) of the Medical Devices Act provides that “no person” may “sell, lease, transfer or use” a medical device that has not been approved, and the only exception is display purposes. Because the subject is “no person”, a medical institution that used it in a procedure is covered too. What a patient can do is simple — ask for the product's approval number, look that number up on the MFDS site, and check that “Export only” reads “No”. That said, a search returning nothing does not mean the product is illegal — brand names generally do not come up at all.

Export-only products receive approval numbers too — we looked them up

Here are two products approved under the same item name (tissue-filling biomaterial, class 4), side by side. This is the result of looking them up directly on 4 September 2026 in the MFDS medical device information portal (emedi.mfds.go.kr, known in Korean as 의료기기안심책방).

The same format of approval number, a different standing
FieldJe-heo 26-667Je-heo 26-657
Item name · classTissue-filling biomaterial · class 4Tissue-filling biomaterial · class 4
Item statusNormalNormal
Standard code (UDI)10
Export onlyNoYes

You cannot tell from the number alone. Both are “Je-heo 26-6xx” and both have an item status of “Normal”. We wrote in the two preceding pieces that you should check the approval number, but the mere fact that a number exists does not tell you whether the product can be used in Korea. You have to look at one more field.

As an incidental clue, whether a standard code (UDI) is attached also differed. The standard code applies to medical devices distributed · sold in Korea, while export-only products may follow the standards of the destination country. But we checked only two, so please use this as a supporting clue only.

Why are there so many export-only items

On the same day we counted the whole set of items returned by a search for “tissue-filling biomaterial”.

Tissue-filling biomaterial items (looked up 2026-09-04)
Total1,010
Export only = Yes745 (73.8%)
Export only = No265 (26.2%)

Three out of every four are export-only. This is not a number produced by anyone doing something wrong; it appears to be the result of how the framework is designed. For a medical device intended solely for export, part of the documentation to be submitted is waived, and the applicant is required to enter “for export only” in the remarks field of the application. A class 4 approval for domestic sale requires clinical data, whereas the burden for export-only is far lighter. From a manufacturer's point of view the cost of entry does not compare.

Put the other way round, it runs like this. An export-only product is a product that has not been reviewed for domestic sale. What was waived is precisely the material that confirms safety · performance. We wrote in the preceding piece that “approval is not a guarantee of efficacy”; for export-only products the threshold of that approval is itself different.

One thing as you read the number — this is a count of items, not of actual distribution. It does not mean that all 745 are circulating in the market. But the very fact that this many export-only items are entered on the Korean approval register is what creates the need for a patient to check.

Can it be used on a patient in Korea — the statute

The sentence of Article 26 (Prohibition of General Acts), paragraph 1 of the Medical Devices Act runs like this.

“No person may repair, sell, lease, transfer or use a medical device that has not obtained approval or certification or has not been notified … nor may any person manufacture, import, repair, store or display such a device for the purpose of selling, leasing, transferring or using it.”

Two things to look at here. First, the subject is “no person”. Not only manufacturers and distributors but the person who used it is included. Second, the only exception in the proviso is “for the purpose of display at a fair · exhibition · trade show and the like”, and there is no exception relating to export.

A product approved “for export only” is a medical device that has not been approved for domestic sale, so using it in Korea falls within “using a medical device that has not been approved” under the article above. The MFDS answers in its FAQ that “an export-only medical device is a product approved solely for export and not for domestic sale, so separate approval · certification must be obtained if domestic distribution is desired”, and it takes the view that displaying · promoting a product at a domestic trade show also constitutes “domestic distribution”.

Penalties
Breach of Article 26(1)Medical Devices Act, Article 51(1)2 — imprisonment for up to 5 years or a fine of up to 50 million won (under paragraph 2 of the same article, imprisonment and a fine may be imposed together)

Some material on the internet still gives the upper limit of the fine as 20 million won. That is the former text. The current text we confirmed is 50 million won. Still, statutes are amended, so this paragraph is as of 4 September 2026, and if you have occasion to cite it, please re-check the current text on the National Law Information Center.

Distinguish three things — unapproved · export-only · revoked

The conclusion that it cannot be used on a patient in Korea is the same for all three, but what you see on the lookup screen is entirely different.

What you see when you look it up
StatusLookup resultWhere to look
UnapprovedNot on the register at all — 0 resultsThere is no number, or it returns nothing
Export-onlyProperly listed · has an approval number · item status “Normal”“Export only” = Yes
Revoked · withdrawn · expiredOn the register, but the status is not normal“Item status”
Usable in KoreaProperly listed · status normalExport only = No

The middle row is the most confusing one. An export-only product looks entirely unproblematic — it has a proper number and its item status is “Normal” too. It only comes out when you open the detailed information and look at one more field.

How to look it up yourself

Anyone can do it on the MFDS medical device information portal (emedi.mfds.go.kr). No account is needed.

The lookup procedure
1Ask the clinic for the product's marketing approval number (Je-heo · Su-heo · Je-in · Su-in · Je-sin · Su-sin + a number)
2On the portal, go to Item and company search
3Enter it in the approval number field with no spaces — for example “제허26-657호”
4Click the result row to open the detailed information
5In the product information table look at three things — item status (is it normal?) · Export only (is it No?) · company name · country of manufacture (does it match what you were told?)

If you were not given an approval number, you can search by model name, and there is also a function that looks the product up from a photograph of the barcode on the box. Opening [Expand search conditions] on the search screen lets you filter on “Export only” itself (though it has to be used together with a search term).

“It does not come up” does not mean it is illegal

This is a trap we ran into ourselves. Brand names generally do not come up. We put well-known filler brand names into the name field and into the product name field of the expanded conditions, and got 0 results. The reason is that the “product name” field in the detailed information is empty for many items — the domestic product we looked at above had a blank product name too. So a brand name not coming up is a common thing and proves nothing in itself. Always look it up by approval number.

There is one more limit to the lookup. What the lookup tells you goes as far as “whether the product with that approval number can be used in Korea”. Whether what is actually in the syringe in front of you is really that product is not something a lookup can tell you. There is also the option of reading the Korean labelling, but if the clinic opens the original packaging and uses it, the patient has no chance to see it. This is a fundamental limit of the method, and it is why we do not write “check it and you can relax”.

The differences visible on the packaging

Under Article 20 of the Medical Devices Act, there are things that must be written on the container or outer packaging of a product distributed domestically.

What a domestic product must carry / what an export-only product need not
ItemDomesticExport-only
Korean-language labellingIn principle KoreanThe language of the destination country is allowed
Approval · certification · notification numberRequiredMay follow the destination country's standards
Manufacturer · importer, and for imports the original manufacturerRequiredDestination country's standards
Lot number · date of manufacture (or expiry date where there is one)RequiredDestination country's standards
Standard code (UDI)RequiredDestination country's standards

So if there is no Korean labelling at all, or there is no original manufacturer · Korean approval number · Korean expiry date and everything is printed in a foreign language only, there is a possibility that the product is not labelled for domestic distribution. But this is a clue and not proof — if the clinic opens the box and uses it you cannot see it, and to check you end up back at looking up the approval number.

Parallel imports · buying from overseas

This comes out of the same article. Neither a clinic nor an individual may use in Korea a medical device bought directly from overseas. Importing is something that a person holding an import business licence is to do, and a product that did not come in that way becomes “a medical device that has not been approved”. The same applies even if you bought it for your own use, on the MFDS's interpretation.

Buying a device you like on the spot at a trade show is not allowed either. As we saw above, the only exception in Article 26(1) is “for the purpose of display”, not sale or purchase.

In March 2025 the MFDS announced that, together with consumer groups · associations, it would build a standing monitoring system for the illegal distribution of medical devices bought from overseas. The items mentioned as having been caught were laser hair-removal devices · blood pressure monitors · cupping devices · anti-snoring devices · anti-bruxism devices · nebulisers and the like.

Have there been actual enforcement cases — what we confirmed and what we could not

We will write this honestly. We did not find a published case of enforcement · sanction of the type “a Korean clinic or hospital performed a procedure on a patient with a device approved for export only”.

That does not mean “such a thing does not happen”. There appears to be a structural reason — the MFDS's public notices of administrative sanctions on medical devices are directed at manufacturers · importers · distributors, and a medical institution is a “user” under the Medical Devices Act and so is not a subject of those sanctions. Sanctions on the medical institution side go down the criminal route, and that is not tallied in a public database. So it is not confirmable from public material is the accurate way to put it.

The judgments we did confirm are somewhat different in character. In a case involving the distribution of laser devices and the like imported without approval, Article 26(1) was applied and a suspended prison sentence and confiscation were handed down. And in a case under the Pharmaceutical Affairs Act, concerning medicines rather than devices, a ruling came in 2024 and 2025 that “where a manufacturer transfers ownership to a domestic exporter for consideration, that constitutes a sale even if the goods are to be exported in future”. Because it is not a medical device case it cannot be carried across as it stands, but the direction — that “even something made for export becomes a domestic sale if it changes hands in Korea” — is worth noting.

Asking is not an unreasonable request

We would rather this piece were not read as “be suspicious of your clinic”. Article 20 of the Medical Devices Act places the duty of writing the approval number on the product container · outer packaging on the manufacturer · importer. If a product came through normal distribution, the clinic is in a position to be able to check that number.

So “please tell me the product's approval number” is not an unreasonable request but an entirely normal verification step. That is particularly so for class 4 products. If you ask us, we will tell you. And we mean to raise it in consultation before you ask.

The two preceding pieces in this series are worth reading alongside — what the approval number in a clinic's material means covers the difference between a marketing approval number and an advertising review number, and Jeheo · suheo · certification covers what a single line of a number can and cannot confirm.

In summary — what is confirmed, what we could not confirm

The grade of the evidence in this piece
CategoryContent
ConfirmedExport-only products also receive a genuine marketing approval number (Je-heo 26-657, looked up for real) · of tissue-filling biomaterial items, 745 of 1,010 (73.8%) are export-only · the subject of Article 26(1) of the Medical Devices Act is “no person” and the only exception is display purposes · the penalty under Article 51(1)2 is imprisonment for up to 5 years or a fine of up to 50 million won · the “Export only” field is public in the detailed information on the MFDS portal · brand names generally do not come up (the product name field is blank for many items)
Could not confirmA published case of enforcement · sanction against a clinic or hospital that used an export-only product on a patient in Korea — this does not mean none exists but that it is not confirmable from public material · whether a separate sanction relating to the medical licence follows the use of an unapproved medical device · a specific report that the domestic and export-only versions of the same brand actually differed in specification — we confirmed only that they may differ as a matter of the regulatory framework
To read with careAmong the judgments cited, the part that “it is a sale if it changes hands in Korea even when destined for export” is a Pharmaceutical Affairs Act · medicines case and not a medical device case · the article number of the notification underlying the export-only route is something we confirmed from the original text of an earlier edition and from secondary material, so it needs re-checking against the current edition · the 745 is a count of items, not of distribution · the statutory content of this piece is as of 4 September 2026

The one line this piece is meant to leave. Having an approval number and being usable in Korea are two different things. Verification is not one line of a number but three fields — the number + the item status + Export only.

Frequently asked questions

Is an export-only medical device an unapproved product?

No. Export-only products receive an MFDS marketing approval number too. The export-only product we looked up had the number “Je-heo 26-657” and its item status was “Normal” as well. The format is exactly the same as the number on a domestic product. It is simply that it has not been reviewed for domestic sale, and so it cannot be used on a patient in Korea.

What is different from a domestic product, then?

The documentation differs. For a product intended solely for export, part of the documentation to be submitted is waived and “for export only” is entered on the application. What is waived is the material that confirms safety · performance. The labelling differs too — a domestic product must in principle carry the approval number · original manufacturer · expiry date · standard code in Korean, while an export-only product may follow the standards of the destination country.

Is it illegal for a clinic to use an export-only product?

Article 26(1) of the Medical Devices Act provides that “no person” may “sell, lease, transfer or use” a medical device that has not been approved, and the only exception is display purposes. An export-only product is not a product approved for domestic sale, so it falls within this. Because the subject is “no person”, a medical institution that used it is covered too, and the penalty is imprisonment for up to 5 years or a fine of up to 50 million won.

How do I check whether the product used on me was a domestic one?

Ask the clinic for the marketing approval number, then enter that number with no spaces in “Item and company search” on the MFDS medical device information portal (emedi.mfds.go.kr). Click the result to open the detailed information and there is an “Export only” field. It has to read “No” for the product to be usable in Korea. Check as well that the item status is “Normal” and that the company name and country of manufacture match what you were told.

I searched by brand name and nothing came up. Is something wrong?

That on its own proves nothing at all. When we checked for ourselves, putting well-known filler brand names into the name field and into the product name field returned 0 results. The reason is that the “product name” field in the detailed information is empty for many items — a perfectly normal domestic product had a blank product name too. Always look it up by approval number.

Would it be rude to ask for the approval number?

Not at all. Article 20 of the Medical Devices Act places the duty of writing the approval number on the product container or outer packaging on the manufacturer · importer. If a product came through normal distribution, the clinic is in a position to be able to check that number. “Please tell me the product's approval number” is an entirely normal verification step, and all the more so for a product classified as class 4, as fillers are.

Are there really that many export-only items?

As of 4 September 2026, of the 1,010 items returned by a search for “tissue-filling biomaterial”, 745 (73.8%) were export-only. Three out of four. That said, this is a count of items entered on the approval register, not actual distribution. It does not mean that all 745 are circulating in the market.

What about a product bought directly from overseas?

Neither a clinic nor an individual may use one. Importing is something a person holding an import business licence is to do, and a product that did not come in that way becomes “a medical device that has not been approved”. The same applies even if you bought it for your own use, on the MFDS's interpretation. Buying on the spot at a trade show is not allowed either — the only exception in the article is “for the purpose of display”, not sale or purchase.

If the lookup comes back normal, can I relax?

That is where this method reaches its limit. What the lookup tells you goes as far as “whether the product with that approval number can be used in Korea”, and whether what is actually in the syringe in front of you is really that product is not something a lookup can tell you. There is the option of reading the Korean labelling, but if the clinic opens the original packaging and uses it there is no chance to see it. That is why we do not write “check it and you can relax”.

Has any clinic actually been caught?

We did not find a published case of that type. That does not mean “such a thing does not happen” — the MFDS public notices of administrative sanctions are directed at manufacturers · importers · distributors, and a medical institution is a “user” under the Medical Devices Act and so is not a subject of those sanctions, which means sanctions on the medical institution side go down the criminal route and are not tallied in a public database. It is not confirmable from public material is the accurate way to put it.

Who wrote this

Written and reviewed by Lee Chi-Hak, MD, medical director of Miso Clinic in Daegu, South Korea. Every study cited above is given together with its design, its size and the limitations the authors themselves recorded, and where we could not find data, we have said that we could not find any.

Miso Clinic
Medical directorLee Chi-Hak, MD
Address4F Bombom Building, 125 Dongdeok-ro, Jung-gu, Daegu, South Korea · Exit 1, Kyungpook National University Hospital Station
Phone+82-53-428-2700
HoursWeekdays 11:00–19:00 (lunch 13:00–14:00) / Saturday 10:00–16:00 (no lunch break) / Closed Sundays and public holidays
ColumnsAll clinical columns
Reference libraryAll booster and device references

References

  1. The actual lookup results (4 September 2026, the MFDS medical device information portal) — Je-heo 26-667 (domestic, standard code 1, Export only “No”) and Je-heo 26-657 (export-only, standard code 0, Export only “Yes”). Both are item name “tissue-filling biomaterial”, class 4, item status “Normal”. The presence or absence of a standard code is something we observed in two records, so please use it as a supporting clue only.
  2. The item count (looked up the same day) — name “tissue-filling biomaterial”, 1,010 in total, Export only “Yes” 745, “No” 265. 745 + 265 = 1,010, so the total adds up. These are item counts, not distribution volumes.
  3. The Medical Devices Act, Article 26 (Prohibition of General Acts), paragraph 1 — “No person may repair · sell · lease · transfer or use a medical device that has not obtained approval or certification or has not been notified …”. The only exception in the proviso is “for the purpose of display at a fair · exhibition · trade show and the like”, and there is no export-related exception.
  4. The Medical Devices Act, Article 51(1)2 — a breach of Article 26(1) carries imprisonment for up to 5 years or a fine of up to 50 million won, with imprisonment and a fine imposable together under paragraph 2 of the same article. The “20 million won” circulating on the internet is the former text.
  5. The Medical Devices Act, Article 20 (Matters to be Stated on Containers and the Like) — the trade name and address of the manufacturer · importer, the original manufacturer for imported goods, the approval (certification · notification) number and the name, the lot number and date of manufacture (or the expiry date where there is one), the weight or unit of packaging, the words “medical device”, a single-use marking, the medical device standard code, and the fact that the package insert is provided online together with the address. Article 23 requires that these be stated accurately in Korean.
  6. Enforcement Rule of the Medical Devices Act, Article 42(2) — an export-only medical device on which the required matters are stated in accordance with the standards of the destination country is an exception to the Article 20 requirements. Article 44(1)1(a) — an export-only medical device on which the statements are in the language of the destination country is an exception to the Korean-language principle.
  7. The basis of the export-only route — the Medical Devices Act itself has no separate article on “export-only medical devices”; the waiver of submitted documents and the entry of “for export only” in the remarks field are set out in the Regulation on the Approval · Notification · Review of Medical Devices (an MFDS notification). We confirmed the current notification number, but the article number is something we confirmed from the original text of an earlier edition and from secondary material, so it needs re-checking against the current edition.
  8. MFDS interpretation — the frequently asked questions collection (medical device field). “An export-only medical device is a product approved solely for export and not for domestic sale, so separate manufacturing · import approval · certification must be obtained if domestic distribution is desired”. Displaying · promoting at a domestic trade show also “constitutes domestic distribution”.
  9. Standard code (UDI) — guidance from the standard code administering body: the standard code applies to medical devices distributed or sold in Korea, and for export-only medical devices it is sufficient to follow the labelling rules required by the destination country.
  10. Judgment — Uijeongbu District Court 2019 Godan 5958 (decided 26 May 2021). In a case of unapproved import · illicit import, the former Medical Devices Act Article 51(1) · Article 26(1) and others were applied, giving 1 year 6 months' imprisonment · 3 years' suspension · confiscation of 87,581,400 won. This case is of the unapproved-import type, not “domestic use of an export-only product”.
  11. Reference judgments (a different statute) — Daejeon District Court 2022 Guhap 106940 (decided 12 June 2024) and Daegu District Court 2022 Guhap 24659. These are Pharmaceutical Affairs Act · medicines cases, and the ruling was that “where a manufacturer transfers ownership to a domestic exporter for consideration, that constitutes a sale even if the goods are to be exported in future”. Because they are not medical device cases they cannot be applied as they stand, and we have taken only the direction from them.
  12. Buying from overseas · parallel imports — in March 2025 the MFDS announced, jointly with consumer groups · associations, that it would build a standing monitoring system for the illegal distribution of medical devices bought from overseas. Neither a clinic nor an individual may use in Korea a product that did not come through a proper import route, and the same applies even for personal use, on the MFDS's interpretation.
  13. What we could not confirm — (1) a published case of enforcement · sanction against a clinic or hospital that used an export-only product on a patient in Korea (this does not mean none exists; the MFDS public notices of administrative sanctions are directed at manufacturers · importers · distributors, and because a medical institution is a “user” it is not among them, so such cases are not tallied in public material) (2) whether a separate sanction relating to the medical licence attaches to the use of an unapproved medical device (3) a specific report that the domestic and export-only versions of the same brand actually differed in specification (we confirmed only that they may differ as a matter of the regulatory framework) (4) an exhaustive search of the MFDS administrative sanctions board.
  14. The statutory content of this piece is as of 4 September 2026. Statutes and notifications are amended, so if you have occasion to cite them, please re-check the current text on the National Law Information Center.

Everything in this column is general information and does not replace medical diagnosis or treatment. Effects and side effects vary with individual skin condition, age and underlying illness, and the same result is not guaranteed for everyone. Any decision to proceed should be made in an in-person consultation with a physician.